Skip to main content
Topic: Lockmass gap correction (Read 15051 times) previous topic - next topic

Re: Lockmass gap correction

Reply #15
Yes of course.
Please send me a PM with your email address and I'll send you a link to a shared dropbox folder

(here is my email : raphael.bilgraer@parisdescartes.fr)

Thanks.

Raphaël

Re: Lockmass gap correction

Reply #16
Ok I found the problem. The line in the raw data I used to detect if an MS1 function was there have been changed with the new masslynx apparently. I really wonder how it could have worked with your other data.

So now my problem is that I don't see a really good way to detect MS1 anymore (as oppose to MS2). So I set it to assume that if there is no MSn then there is MS1.
I could probably find a better solution but I don't have time ATM to wrap my head around the very convoluted data waters writes.

The script below should work for you. Otherwise get back to me. Please check the output against what masslynx show just to verify that it is doing what it is suppose to.


Jan.


Code: [Select]
##############################################
## Settings (dirs have to use "/", not "")
##############################################
indir = 'C:/Users/jans/Desktop/ICHTYOSE_2013_06_07_2590'


outdir = 'C:/Users/jans/Desktop/ICHTYOSE_2013_06_07_2590/converted'
##############################################




##############################################
## Libraries
##############################################
library(xcms)
library(MetShot)
library(stringr)
##############################################




##############################################
## Functions
##############################################
no_scan_empty_warn <- function(w) if( any( grepl( "MS1 scans empty. Skipping profile matrix calculation.", w) ) )
  invokeRestart( "muffleWarning" )

no_split_error_warn<- function(w) if( any( grepl( "MSn information will be dropped", w) ) )
  invokeRestart( "muffleWarning" )


remakeTIC<-function(object){
  object@tic<-numeric(length(object@scanindex))
  for(i in 1:length(object@scanindex)){
    object@tic[i]<-sum(getScan(object, i)[,"intensity"])
  }
  return(object)
}



LockMassFun_rem<-function(object,idx){
  #idx<-AutoLockMass(object) 
  #idx=FindLockMass(object)
  #select=c(0,1)
 
  if(missing(idx)){stop('Parameters are missing')}
 
  select=c(1,1)
 
  #  if (any(diff(idx)==1)){
  #    select=c(0,1)
  #  }else{
  #    select=c(1,1)
  #  }
 
 
 
  idx.TF<-as.logical(rep(select, length(idx)/2))
  idx.TF<-which(idx.TF != 0)
  inx<-rep(TRUE, length(object@scanindex))
  inx[idx[idx.TF]]<-FALSE
 
 
  # separate scans into list
  scans=list()
  for (i in 1:length(object@scanindex)){
    scans[[i]]=getScan(object,i) 
  }
 
 
  # Just removed unwanted scans...
  scans=scans[inx]
 
 
  # Create new scanindex
  scanindex_new=numeric()
  for (i in 1:length(scans)){
    if (i==1){
      scanindex_new[i]=0
    }else{
      scanindex_new[i] = scanindex_new[i-1] + nrow(scans[[(i-1)]])
    }
  }
 
 
  # Put data into new object
  ob<-new("xcmsRaw")
  ob@env = new.env(parent=.GlobalEnv)
 
  scans=do.call(rbind,scans)
  ob@env$mz<-scans[,'mz']
  ob@env$intensity<-scans[,'intensity']
 
  ob@scantime <-object@scantime[inx]
  ob@scanindex <-as.integer(scanindex_new)
  ob@polarity  <- object@polarity[inx]
  ob@acquisitionNum <- 1:length(ob@scanindex)
  ob@profmethod    <- object@profmethod
  ob@mzrange        <- object@mzrange
  ob@filepath      <- object@filepath
  profStep(ob) =  profStep(object)
  ob=remakeTIC(ob)
 
  return(ob)
}







LockMassFun_rep<-function(object,idx){
  #idx<-AutoLockMass(object)
  #idx=FindLockMass(object)
  #select=c(0,1)
 
 
  if(missing(idx)){stop('Parameters are missing')}
  select=c(1,1)
 
  #  if (any(diff(idx)==1)){
  #    select=c(0,1)
  #  }else{
  #    select=c(1,1)
  #  }
 
  idx=idx[idx>1] # We don't allow to replace scan 1 since there is nothing before to replace with
 
  idx.TF<-as.logical(rep(select, length(idx)/2))
  idx.TF<-which(idx.TF != 0)
  inx<-rep(TRUE, length(object@scanindex))
  inx[idx[idx.TF]]<-FALSE
 
 
  # separate scans into list
  scans=list()
  for (i in 1:length(object@scanindex)){
    scans[[i]]=getScan(object,i) 
  }
 
 
  # Here we do the fillin instead
  for (i in idx[idx.TF]){
    scans[[i]]=scans[[i-1]]
  }
 
 
  # Create new scanindex
  scanindex_new=numeric()
  for (i in 1:length(scans)){
    if (i==1){
      scanindex_new[i]=0
    }else{
      scanindex_new[i] = scanindex_new[i-1] + nrow(scans[[(i-1)]])
    }
  }
 
 
  # Put data into new object
  ob<-new("xcmsRaw")
  ob@env = new.env(parent=.GlobalEnv)
 
  scans=do.call(rbind,scans)
  ob@env$mz<-scans[,'mz']
  ob@env$intensity<-scans[,'intensity']
 
  ob@scantime      <-object@scantime
  ob@scanindex      <-as.integer(scanindex_new)
  ob@polarity      <- object@polarity
  ob@acquisitionNum <- object@acquisitionNum
  ob@profmethod    <- object@profmethod
  ob@mzrange        <- object@mzrange
  ob@filepath      <- object@filepath
  profStep(ob)      <- profStep(object)
  ob=remakeTIC(ob)
 
  return(ob)
}



MSnSplit<-function(object,f){
  ob = withCallingHandlers( split(msn2xcms(object),  f  )$'TRUE' , warning = no_split_error_warn )
 
  ob@env$msnIntensity    =    ob@env$intensity
  ob@env$msnMz          =    ob@env$mz
 
  ob@tic                      = numeric()
  ob@msnScanindex            = ob@scanindex
  ob@scanindex                = integer()
  ob@scantime                = numeric()
  ob@acquisitionNum          = integer()
  ob@mzrange                  = numeric()
  ob@msnAcquisitionNum        = ob@msnAcquisitionNum[f]
  ob@msnPrecursorScan        = ob@msnPrecursorScan[f]
  ob@msnLevel                = ob@msnLevel[f]
  ob@msnRt                    = ob@msnRt[f]
  ob@msnPrecursorMz          = ob@msnPrecursorMz[f]
  ob@msnPrecursorIntensity    = ob@msnPrecursorIntensity[f]
  ob@msnPrecursorCharge      = ob@msnPrecursorCharge[f]
  ob@msnCollisionEnergy      = ob@msnCollisionEnergy[f]
 
  return(ob)
}




MSnExtractLockmass<-function(object,MSparam=NULL){
  if (missing(MSparam)){stop('MSparam has to be set.')}
 
  lockmass=grep('Lock Mass',MSparam)
 
  if (!(length(lockmass)==0)){
  lockmass=MSparam[lockmass[length(lockmass)-1]]
  lockmass = strsplit(lockmass,"t")
  lockmass = as.numeric(lockmass[[1]][length(lockmass[[1]])])
 
  }else{
    lockmass=grep('EDC Mass',MSparam)
    lockmass=MSparam[lockmass]
    lockmass = strsplit(lockmass,"t")
    lockmass = as.numeric(lockmass[[1]][length(lockmass[[1]])])
  }
 
 
  lockhist_freq = as.matrix(table(object@msnPrecursorMz))
  lockhist_masses = as.numeric(rownames(lockhist_freq))
  lockhist_masses = lockhist_masses[  which.min(abs(lockhist_masses-lockmass))  ]
 
  return(lockhist_masses)
}
##############################################



##############################################
## Get file list
##############################################
files = list.files(indir, pattern='.raw', full.names=T)

namelist=character()
filenames=character()

log_convert = list()
log_warnings=character()
log_convert_warning = list()
log_convert_warning[[1]]=' '
##############################################





for (i in 1:length(files)) {
 
  ##############################################
  ## Check if lockspray is present. TODO: Scream at waters
  ##############################################   
  input_method = scan(paste(files[i],'/_extern.inf',sep=""), character(0), sep = "n",quiet=T)
  lockspray_present = any(grepl('Lock Spray Configuration',input_method))
 
  if (!lockspray_present){
    warning(paste('LockMass seems to be disabled for file ',basename(files[i]),'!', ' Saving data as is.',sep=""),immediate. = T)
  }
  ##############################################
 
 
  ##############################################
  ## Get name of sample
  ##############################################
  name = scan(paste(files[i],'/_HEADER.TXT',sep=""), character(0), sep = "n",quiet=T)
  name_where = grepl('Sample Description',name)
  namelist[i] = strsplit(name[name_where],': ')[[1]][2]
  rm(name,name_where)
  ##############################################
 
 
 
  ##############################################
  ## Get polarity
  ##############################################
  ionmode = scan(paste(files[i],'/_extern.inf',sep=""), character(0), sep = "n",quiet=T)
 
 
  if (!(any(grepl('ES+',ionmode,fixed = T)) | any(grepl('ES-',ionmode,fixed = T)))){warning('Ionization mode cannot be detected!')
                                                                                   
  }else{
    if(any(grepl('ES+',ionmode,fixed = T))){ionmode='positive'
    }else{ionmode='negative'}
  }
  ##############################################
 
 
 
 
  ##############################################
  ## Get the functions and figure if there is MS1, MSn and/or MSE
  ##############################################
  func_start = grep('Function Parameters - Function',input_method,fixed=T)
  func_end  = c(  grep('Function Parameters',input_method,fixed=T)-1  ,  length(input_method) )
  func_end = func_end[  2:length(func_end)  ]
 
 
  MSE=F
  MSn=F
  MS1=F
 
  functions=list()
  for (i2 in 1:(length(func_start))){
    functions[[i2]]=input_method[    func_start[i2]:func_end[i2]    ] 
  }
 
 
  not_ref = unlist(lapply(functions, function(x) !grepl(' - REFERENCE',x[1])))
  MSn = any(unlist(lapply(functions, function(x) grepl('TOF MSMS FUNCTION',x[1],fixed=T))))
 
 # MS1 = any(unlist(lapply(functions, function(x) grepl('TOF MS FUNCTION',x[1],fixed=T))))
#  MS1 = any(unlist(lapply(functions, function(x) grepl('Trap MS Collision Energy (eV)ttt4.0',x,fixed=T))))
  if (!MSn){MS1=T}
 
 
  if (!MSn){
  for (i2 in which(not_ref)){
    if (any(grepl('Collision Energy Ramp Start',functions[[i2]],fixed=T))){
      MSE=T
    }
  }
  }
  ##############################################
 
 
 
 
  ##############################################
  ## Convert the data
  ##############################################
  tempfile = tempfile(pattern = basename(gsub('.raw','',files[i])), tmpdir = '/', fileext = ".mzXML")
 
 
  # if there is mobility data we need to make a copy and remove it. Otherwise masswolf doesn't work
  if (file.exists(  paste(  files[i],'/apex3D',sep=''))){
    input_file=  paste(  gsub('.raw','',files[i],fixed=T),'_temp.raw',sep='')
    dir.create(input_file)
    to_copy=list.files(path = files[i], all.files = T,full.names = T, recursive = F,ignore.case = T, include.dirs = T,no.. = T)
    to_copy=to_copy[      !(grepl('/_mob',to_copy,fixed=T) | grepl('/apex3D',to_copy,fixed=T) | grepl('.cdt',to_copy,fixed=T) | grepl('.ind',to_copy,fixed=T))  ]
    temp=file.copy(to_copy, input_file,overwrite=F,recursive=T)
    rm(temp)
  }else{
    input_file= files[i]
  }
 
 
  # Convert with mass wolf
  if (MSE){
    log_convert[[i]]=system(paste('masswolf --MSe --mzXML ',""",input_file,""",' ',""",outdir,tempfile,""",sep=""), intern=T)
  }else{
    log_convert[[i]]=system(paste('masswolf --mzXML ',""",input_file,""",' ',""",outdir,tempfile,""",sep=""), intern=T) 
  }
 
  # Delete temporary file of created (for files that include mobility data)
  if (file.exists(  paste(  files[i],'/apex3D',sep=''))){
    unlink(input_file,recursive=T)
  }
   
   
 
  # Read the raw Waters data
  raw_file_data = scan(paste(outdir,tempfile,sep=''), character(0), sep = "n",quiet=T)
 
  scan_start = grep('<scan num=',raw_file_data,fixed=T)
  scan_end = c(    scan_start[2:length(scan_start)]-1    ,  length(raw_file_data)  )
 
 
  raw_file_data_scans=list()
  for (i2 in 1:(length(scan_start))){
    raw_file_data_scans[[i2]]=raw_file_data[    scan_start[i2]:scan_end[i2]    ]
   
  }
  ##############################################
 
 
 
  ##############################################
  ## Read retention times and scan numbers in original data
  ##############################################
  # Get scan retention times
  raw_file_data_ret = lapply(raw_file_data_scans, function(x) as.numeric(str_trim(str_replace(str_replace(x[    grep('retentionTime',x)  ],'retentionTime="PT', ''),'S"',''),side='both'))  )
  raw_file_data_ret = unlist(raw_file_data_ret)
 
  # Get which are lockmass (calibration) scans
  is_lockmass_scan =  lapply(raw_file_data_scans, function(x)  any(grepl('scanType="calibration"',x,fixed=T))  )
  is_lockmass_scan = unlist(is_lockmass_scan)
 
 
  # Get which are MSE or MS2 scans
  #is_MSE_or_MSn_scan =  lapply(raw_file_data_scans, function(x)  any(grepl('msLevel="2"',x,fixed=T))  )
  #is_MSE_or_MSn_scan = unlist(is_MSE_scan)
  ##############################################
 
 
 
  ##############################################
  ## Read file
  ##############################################
  xr_org = withCallingHandlers( xcmsRaw(paste(outdir,tempfile,sep=""),includeMSn=T), warning = no_scan_empty_warn )
  ##############################################
 
 
 
  ##############################################
  ## if it is an MSE file, write seperate MSE file
  ##############################################
  if (MSE){
    xr_MSE =new('xcmsRaw')
   
    xr_MSE@env = xr_org@env
    xr_MSE@msnScanindex = xr_org@msnScanindex
    xr_MSE@msnAcquisitionNum = xr_org@msnAcquisitionNum
    xr_MSE@msnPrecursorScan = xr_org@msnPrecursorScan
    xr_MSE@msnLevel = xr_org@msnLevel
    xr_MSE@msnRt = xr_org@msnRt
   
    xr_MSE@msnPrecursorMz = xr_org@msnPrecursorMz
    xr_MSE@msnPrecursorIntensity = xr_org@msnPrecursorIntensity
    xr_MSE@msnPrecursorCharge = xr_org@msnPrecursorCharge
    xr_MSE@msnCollisionEnergy = xr_org@msnCollisionEnergy
    xr_MSE@filepath = xr_org@filepath
   
   
    filename=basename(gsub('.raw','_MSE.mzData',files[i]))
    write.mzdata(xr_MSE,paste(outdir,'/',filename,sep=""))
  }
  ##############################################
 
 
 
  ##############################################
  ## Write MS1 data
  ##############################################
  if (MS1){
   
   
  if (lockspray_present){
         
      idx=match(raw_file_data_ret[is_lockmass_scan],xr_org@scantime)
      idx = idx[!is.na(idx)]
     
      xr_fixed=LockMassFun_rep(xr_org,idx)
      xr_fixed@polarity=factor(rep(ionmode,1,length(xr_fixed@polarity)),levels=c('negative','positive','unknown'))
     
    }else{
      xr_fixed=xr_org
      xr_fixed@polarity=factor(rep(ionmode,1,length(xr_fixed@polarity)),levels=c('negative','positive','unknown'))
    }
 
 
 
 
  # Write fixed file
  filenames[i]=basename(gsub('.raw','.mzData',files[i]))
  write.mzdata(xr_fixed,paste(outdir,'/',filenames[i],sep=""))
 
  temp1 = plotChrom(xr_org)
  temp2 = plotChrom(xr_fixed)
  }
  ##############################################
 
 
 
 
 
  ##############################################
  ## Write MSn data
  ##############################################
  if (MSn){
   
   
    if (lockspray_present){
     
      LockMass = MSnExtractLockmass(xr_org,MSparam=input_method)
      f = xr_org@msnPrecursorMz != LockMass
     
      if(any(f)){
        log_convert_warning[length(log_convert_warning)+1]=print(paste('Lockmass detected at ',formatC(LockMass,digits=7),'Da in range ',format(min(xr_org@msnRt[f])/60,digits=2),'min to ',format(max(xr_org@msnRt[f])/60,digits=2),'min for file: ',basename(files[i]),sep=""))
        xr_fixed = MSnSplit(xr_org,f)
        xr_fixed@polarity=factor(rep(ionmode,1,length(xr_fixed@polarity)),levels=c('negative','positive','unknown'))
      }else{
        log_convert_warning[length(log_convert_warning)+1]=print(paste('Lockmass was enabled but no lockmass scans could be found. Data written as is for file: ',basename(files[i]),sep=""))
        xr_fixed=xr_org
      }
   
     
    }else{
      xr_fixed=xr_org
      xr_fixed@polarity=factor(rep(ionmode,1,length(xr_fixed@polarity)),levels=c('negative','positive','unknown'))
    }
   
   
    # Write fixed file
    filenames[i]=basename(gsub('.raw','.mzData',files[i]))
    write.mzdata(xr_fixed,paste(outdir,'/',filenames[i],sep=""))
   
    temp1=msn2xcms(xr_org)
    temp1=remakeTIC(temp1)   
    temp1=plotTIC(temp1)
    temp2=msn2xcms(xr_fixed)
    temp2=remakeTIC(temp2)   
    temp2=plotTIC(temp2)
  }
  ##############################################
 
 
 
 
  ##############################################
  ## Debugging plots
  ##############################################
  ## debugging, plotting
  dir.create(paste(outdir,'/debug/',sep=""),showWarnings=F)
  png(paste(outdir,'/debug/',basename(gsub('.raw','.png',files[i])),sep=""),width=2000,height=1500,res=300)
  plot(temp1,type='l',col='red',xlim=c(-10,max(temp1[,1])))
  lines(temp2,col='green')
  dev.off()
  rm(temp1,temp2)
 
 
 
  # delete temp file
  unlink(paste(outdir,tempfile,sep=""))
  rm(xr_org,xr_fixed)
  gc()
  log_warnings[i]=warnings()
}
##############################################




write.csv(cbind(filenames,namelist),paste(outdir,'/','filelist.csv',sep=""))


fileConn<-file(paste(outdir,'/','log_convert.log',sep=""))
writeLines(unlist(log_convert), fileConn)
close(fileConn)

fileConn<-file(paste(outdir,'/','log_warnings.log',sep=""))
writeLines(unlist(as.character(log_warnings)), fileConn)
close(fileConn)

fileConn<-file(paste(outdir,'/','log_convert_warning.log',sep=""))
writeLines(unlist(log_convert_warning), fileConn)
close(fileConn)
Blog: stanstrup.github.io

Re: Lockmass gap correction

Reply #17
Jan,

The script works now. Everything seems to be fine !
I'll double check the output and keep you updated if there are some errors.

Thanks a lot.
Best,

Raphaël