MS-DIAL / Re: How to Specify Column Polarity and use Retention Index for ID
Last post by James -I don't know of a way to do this within MS-DIAL, so I would untick the 'use retention information for scoring' box in the Identification tab of the parameter settings, and then compare the calculated retention indices with your putatively identified compound RIs manually.
If there are so many compounds that it isn't feasible or efficient to do it manually, the webchem R package has functions for searching online databases for RI information of the compounds in your dataset and allows you to specify column polarities and retention index types - meaning you could specify only polar Kovats retention indices and then quickly screen for big differences.
Cheers,
James