Hi All,
I was wondering if there is an existing processing/ post-processing solution for side/ partial peak XCMS artifacts.
The first concerns m/z traces that are typically different by 0.01-0.1 Daltons from the main peak and usually within 0.2 minutes (~12s). Example: Images 1 & 2 below.
The second concerns m/z traces often of equivalent m/z but within 0.3 minutes/ 20s from a particularly wide peak. Increasing my upper limit peakwidth did not seem to fix the issue. Maybe bw is too small? Example: images 3 & 4.
From reading previous forums on related topics, increasing mzdiff to as much as 0.1 helps eliminate overlapping mass trace peaks (at the expense of overlapping and poorly resolved isomer peaks), but side peaks of the nature described above remained. Playing with the prefilter option does help eliminate a lot of these, but again some still remain. The instrument used is an Agilent Q-TOF so I'm not sure if the discussion on Orbitrap shoulder peaks applies (though one forum from many years back mentioned that Agilent peaks have problems with ringing).
Basic details:
R version 3.5.3
XCMS version 3.4.4
mset <- xcmsSet("NEG_10",
method = "centWave",
peakwidth = c(5, 40),
ppm = 30,
noise = 100,
snthresh = 8,
mzdiff = 0.05,
prefilter = c(5, 500),
mzCenterFun = "wMean",
integrate = 2,
fitgauss = TRUE,
verbose.columns = FALSE
)
mset2 <- retcor(mset,
method = "obiwarp",
profStep = 1,
center = 5)
mset2 <- group(
mset2,
method = "density",
bw = 3,
mzwid = 0.05,
minfrac = 0.5,
minsamp = 1,
max = 50)
mset3 <- fillPeaks(mset2)
[attachment deleted by admin]
For the orbitrap like shoulder peaks I wrote a filter, xcmsRaw.orbifilter, you can find here https://gitlab.com/R_packages/chemhelper.
It runs through each scan in a file and starts with the largest mass and eliminates everything in a set mz range around that peak that is smaller than some fraction of the main peak.
It runs through all peaks in the scan until all peaks have been "evaluated".
Currently you'd need to do this on all the raw files and write them out to a new set of raw files.
As for the chromatographic side peaks I am not sure you can fix that with parameters (apart from the grouping you already mentioned). Also in your plots they look like to me to be legitimate additional small peaks so I don't see what the peak picker should be doing differently. You have fronting and tailing peaks and I guess that will always be problematic.
Your ~447 peak is very noisy. If your data in general looks like that the matchedfilter algorithm could give you better results. It generally is more robust to noisy data.