The main purpose of the peakwidth parameter is to roughly estimate the peak width range, this parameter is not a threshold.
The wavelets used for peak detection are calculated from this parameter.
If you use HPLC and your peaks are normally 20 - 60 s wide (base peak with), just go with that, i.e. peakwidth=c(20,60)
centWave will still detect peaks that are 15s or 80 s wide!
Important: Do not choose the minimum peak width too small, it will not increase sensitivity, but cause peaks to be split.
Example: peak width ~ 45 s
using peakwidth = c(10,120) the peak will be split in three peaks, each detected as a ~10s wide separate peak (since they are separated by a local minimum) :
[attachment=1:toypq4xs]peakwidth_10_120.png[/attachment:toypq4xs]
using peakwidth = c(20,120) will keep the peak intact :
[attachment=0:toypq4xs]peakwidth_20_120.png[/attachment:toypq4xs]