MS-DIAL v5.1.230807: LOWESS unavailable when trying to normalise August 18, 2023, 01:30:57 AM Hi,I am currently experiencing an issue with MS-DIAL v5.1.230807. After processing my SWATH data, including peak detection, deconvolution, identification, and alignment, I have encountered a problem with the normalization step using LOWESS. The "Normalize" button appears to be greyed out and inaccessible.I want to note that I have QCs available and that I have configured the settings for the overall data, including analytical order and sample type. Despite this, the "Normalize" button remains inactive.Has anyone else encountered a similar issue and found a solution? Your insights would be greatly appreciated.Thank you.Thomas 2 Likes
Re: MS-DIAL v5.1.230807: LOWESS unavailable when trying to normalise Reply #1 – August 30, 2023, 04:56:09 AM We do face the same problem (X500R SWATH, MS-DIAL v5.1.230807, same procedure) and did not find a solution yet, unfortunately. 2 Likes
Re: MS-DIAL v5.1.230807: LOWESS unavailable when trying to normalise Reply #2 – September 04, 2023, 02:33:49 AM Hi Thomas,it seems like we found a solution to the issue:To circumvent the problem, make sure that your injection order shows a QC sample at the first and last position of your batch. If you include different batches in one MSDial file, it needs to be true for all the batches individually.Best,Stefan 1 Likes